  • Unable to calibrate grill ?

    Posted by guest on December 3, 2015 at 12:00 am

    Complaint was uneven heat on 3 burner grill m# sg-36 S# 951393634 south bend. Jb style valves, when set on 300*f  Left is 355 middle 400 right is 300. left and middle have a standby flame adjustment. Right valve does not. adjusted standby flame and checked gas pressures. when all off 6 in wc and a drop of 2 per burner when fully on.  I suspect a faulty pressure regulator  but how much drop should there be when all open.  And what is the best way to calibrate this style.

    ectofix replied 10 years, 11 months ago 1 Member · 2 Replies
  • 2 Replies
  • ectofix

    Member
    December 4, 2015 at 11:53 am

    Don’t know why that 3rd thermostat wouldn’t have a bypass.  All thermostats should be identical.

     

    Regarding the pressure drop, if the gas supply plumbing is adequate for the unit’s (and kitchen’s) Btu demands…and the regulator is functioning properly, then an “ideal” pressure drop should be negligible.  However, I’ve seen up to an inch of WC drop and had to call it good so as not to tell then their entire kitchen needs new gas plumbing.  That 2″WC is too much though.  If you’re a 2″ drop PER THERMOSTAT, then YES, replace the regulator.

     

    That’s all I’ll type on my smartphone.  More about calibrating a griddle later…

  • ectofix

    Member
    December 4, 2015 at 6:24 pm

    Regarding calibration, the instructions are in on pg 23 of this PDF (pg 3 of the SERVICE section):

     

    http://download.partstown.com/is-bin/intershop.static/WFS/Reedy-PartsTown-Site/-/en_US/manuals/SOU-SG-24-E_spm.pdf

     

    Make sure you have the bypass flame set before attempting calibration.  I suggest verifying proper gas pressure and uniformed (as possible) burner flames before proceeding.  A minimum of 7″WC of pressure feeding your regulator.  The manual specifies 4″WC manifold pressure (after the regulator).  For the flames, make sure the orifices are clean, the primary air dampers are adjusted for proper flame characteristics and the burners are clean, in good condition and properly installed.  Also, make sure the flue is completely clear.

     

    Common practice:

     

    If a thermostat is off by 50°F, replace it. If it’s off by that much so suddenly, then something went wrong in it. If a critical part is questionable (such as a thermostat) isn’t OEM, replace it with an OEM part. I strongly advise staying with OEM parts.  This isn’t HVAC or refrigeration – where it’s common practice to match a part as close as you can from something the supply house has.  You’re FAR better off using the part an EQUIPMENT (end-item) manufacturer specifies. When calibrating a griddle thermostat, take your temperature readings just above where the thermostat bulb mounts to the underside. Bear in mind that it’s impossible for the entire griddle surface to be at a uniformed temperature, given the burners’ typical horse-shoe design.  There will ALWAYS be hotter and colder areas.  Temperature consistency only improves when a thicker griddle plate is used in the design. Use a QUALITY thermometer and probe. I personally use a Cooper-Adkins digital meter with their anvil griddle probe.  I have NO faith whatsoever in those dial-type thermometers that are all-in-one sitting there on the griddle (others may disagree).

     

    That oddball, right-hand thermostat with no bypass doesn’t sound like it’s OEM.  But…it’s in there and happens to be working properly, so let it ride.  If it has no bypass adjustment, then it sounds like a snap-action (all-or-nothing) thermostat.  Maybe a GS thermostat.  Our Vulcan V-Series griddles use those.  I don’t like them.

     

    By the way, that is a BJ (not JB) thermostat (short for BJWA) – a RobertShaw throttle-action (modulating) thermostat.

     

    When you’re calibrating a griddle, be v-e-r-y patient.  Watching its temperature swings is like watching grass grow.  So, proper calibration can take awhile – especially if starting from a cold griddle.

     

    As you’re watching those temperature swings, write down the high and low readings for each thermostat.  The average of those two numbers is where it’s presently at.  So, if the low reading is 290° and the high reading is 308°, then the average is darned close to your desired set point.  If it’s off by 20° or more, follow the manual’s instructions (push in the screw, but don’t turn it, turn the knob to what your actual reading is).  After you’ve tweaked the calibration screw, give it a gentle wiggle to be sure it’s locked into place in the little cogs in there.

     

    Check your work.  Let it cycle a time or two to verify that you’re on the mark and everyone is happy.

    –

    –

    –

    That’s all I’ve got to say about that.